I am automating control of a LeCroy oscilloscope using IEEE488.2 and Python/PyVisa.

It seems like it should be relatively straightforward. I use the WF? command to get the waveform + header, I use the Template? command to get the template for interpreting the header, and then I convert the proper 4 byte fields to floats to get the VERTICAL GAIN and VERTICAL OFFSET.

The problem is that this gives me a scaling factor that doesn't match what I see on the oscilloscope screen. In this link I have a picture of the scope screen, an xlsx with the Template/Header/Waveform responses, and the Python code I used to put those responses in the xlsx -> https://drive.google.com/file/d/1FGZZ_1fbPYiSIDmTUJg3PWhE15W3i1KR/view?usp=sharing

Can someone please point out what I am doing wrong? It seems this should be really easy.

Edited: Including pictures of scope picture and my scaled waveform.Scope PictureScaled Waveform from xlsx

Bobby
  • 31
  • 3

1 Answers1

1

The value is a left justified 12bit Q8 number. I suppose I should have noticed the factor of 4 in the pictures.

Bobby