11

I was just perusing the open-source schematic for an ECG amplifier, found here. The schematic contains what I take to be a reference-voltage buffer, with heavy use of capacitive filtering:enter image description here

From what I understand, using an op amp in unity-gain configuration to drive a large capacitive load (C24) will tend to reduce phase margin and make unwanted oscillation likely. So is there some element here that prevents this? In particular I was wondering if C22 played a role.

davidf2281
  • 220
  • 2
  • 7

5 Answers5

26

The usual situation for most op-amps is that you can "hang" capacitance on the output up to a small value of (say) 100 pF and then, if you increase that capacitance, the op-amp misbehaves and turns into an oscillator. For this device (as with most op-amps), that capacitance is shown to be about 150 pF in the data sheet. Figure 41 shows that with 150 pF connected the output is sustaining oscillation. With 130 pF connected there is overshoot and a damped ringing so, in reality, you shouldn't go greater than 100 pF. I wouldn't call it a rule of thumb because some op-amps will "sing" with 22 pF connected. As always, read the data sheet.

\$\color{blue}{\text{But what if you stuck several uF on the output?}}\$

Chances are is that it won't oscillate either (even though the data sheet doesn't appear to explicitly state this). How can this be possible you might ask or, how would I know that it won't oscillate might be another question. The answer can be found in most data sheets but you have to dig-deep and analyse things in detail.

For instance, take figure 10: -

enter image description here

What might this tell us you might ask. It tells me that the internal resistance of the transistor output stage is about 125 ohms under load. With \$V_{DD}\$ at 5 volts, you should be able to see that the slope of the line is about 1 volt per 8 mA (as per my red additions). That's an output resistance of 125 ohms. Later on\$^1\$ I'll mention the light load scenario (blue scribblings).

Next is the open-loop gain characteristic in figure 33: -

enter image description here

I've taken the liberty of drawing a red-line at unity gain. At unity gain (about 3 MHz) can you see that the phase shift line corresponds to about 150 degrees? This means it is 30 degrees away from becoming an oscillator i.e. 30 degrees away from regular inverting feedback becoming non-inverting (or positive) feedback. This 30 degrees is called "the phase margin" and you want that to be as big as possible in order to reduce overshoot and (god forbid) avoid turning the circuit into an oscillator.

To help see this more clearly I've added simulation results. Here's the first one: -

enter image description here

Adding 100 pF moves the phase margin lower i.e. the 30 degrees margin becomes less. How much less does it become you might ask. The answer begins in figure 10 where I estimated the output resistance of the op-amp and got 125 ohms.

That resistance plus 100 pF forms a low pass filter of cut-off 12.7 MHz. At this frequency, the added phase shift is 45 degrees. But, importantly, at 3 MHz the phase shift is only about 12 degrees. Therefore, with 100 pF connected to the output our phase margin has reduced from 30 degrees to 18 degrees. I've used this tool to help me visualize that phase shift. I could calculate it but why bother when Mr Okawa provides such nice tools.

Again, I've added a simulation result: -

enter image description here

If the capacitance were 150 pF, the cut-off has reduced to 8.4 MHz but, of significance, at 3 MHz, the phase shift is 20 degrees and our phase margin has reduced to only 10 degrees. This suggests that the device won't oscillate but I'm just using typical values and not worst case.

\$\color{blue}{\text{But what if you stuck several uF on the output?}}\$

I haven't answered this yet but here goes.

With 10 nF on the output, this brings a cut-off frequency of 127 kHz with the internal equivalent output resistance of 125 ohms. And, if you looked at the open-loop gain graph you'd see that the phase margin is about 80 degrees so, on the face of it, you might be able to use 10 nF but, at 1 MHz the phase margin is 70 degrees AND the 10 nF and 125 ohm produces an added phase shift of around 82 degrees hence, somewhere a little below 1 MHz the op-amp would sing its heart out in a sustained oscillation.

I've added another sim result to show this (it will oscillate around 500 kHz): -

enter image description here

So, here's the crux of what I've been driving towards: -

When the frequency is low the natural phase margin of the op-amp is always good. For example, at 300 Hz, the phase margin is 120 degrees and adding a 1 uF to the output reduces the margin to about 105 degrees. If we took the frequency up to 10 kHz, the op-amp's natural phase margin is about 92 degrees and the output capacitor reduces this to around 10 degrees.

But the overall gain will have dropped due to the output capacitor. At 10 kHz, the open-loop gain is around 300 (50 dB) but the extra loss of gain brought about by the capacitor is about 20 dB. I still think that it's likely that 1 uF would sustain oscillation but what about 10 uF?

10 uF introduces an extra loss of gain of 50 dB at 10 kHz and the phase margin would become about 2 degrees (this assumes a total added phase shift of 90 degrees from the capacitor and output resistor and it can't be any more than this). I think this might still oscillate. It's borderline.

Added 10 uF simulation showing that it is very borderline: -

enter image description here

With 100 uF added, the op-amp would be clear of any problems because the attenuation of the output capacitor and resistor is 60 dB (at 10 kHz) and oscillation cannot be sustained. In other words, the natural phase margin of the op-amp could never be a value that adding another 90 degrees would produce a net gain greater than 1.

For completion, the 100 uF simulation: -

enter image description here


Blue scribble\$^1\$ - if very lightly loaded, the op-amp's output resistance might be seen as 500 ohms and this, has the same effect as making the capacitance 4 times bigger. The impact of this (when you have swamped the output pin with 10 uF for example) is that the output will very likely be stable but, as you drew more output current, the dynamic resistance drops towards 125 ohm and the circuit might suddenly become unstable. Many Pro EEs will have noticed this type of similar effect from time to time.

Andy aka
  • 434,556
  • 28
  • 351
  • 777
  • Thanks for the very detailed response. Would you also agree with vtolentino's opinion that C22 may provide some additional protection against oscillation by pre-charging the output caps? – davidf2281 Apr 17 '20 at 13:47
  • 1
    No, I wouldn't - you need to ask him how he came to that conclusion. – Andy aka Apr 17 '20 at 13:52
  • Could you speculate on the function of C22? I'm finding it difficult to see its purpose. – davidf2281 Apr 17 '20 at 13:55
  • 1
    I don't see a reason to have C22 at all. Many designs have oddities based on false beliefs and this might be one of them. – Andy aka Apr 17 '20 at 13:58
  • Ha, that makes me feel better! Thanks again. – davidf2281 Apr 17 '20 at 14:01
  • Andy, my understanding of amplifier stability is that oscillation only occurs if, at some frequency in the open loop response, the loop gain is at or close to 1 when the phase is -180 deg. If the loop gain is a fair bit less than 1 when the phase is -180 deg then stability should be guaranteed but loop gain greater than 1 when the phase is -180 deg doesn't guarantee instability, it just means that the amp will be unstable if at some higher frequency the phase is still at -180 degs by the time the loop gain has dropped to unity. Barkhaussen Criteria: Loop Gain = 1, -180 deg for instability. –  Apr 18 '20 at 14:22
  • I may be wrong, but I believe the point of high phase margin is to prevent large overshoots in the output, as well as to ensure that any parasitic poles on the op amp, circuit, PCB, etc. don't tip it over the edge of instability. So the results here make me a little nervous. With phase margins in the single degrees, personally I would take extra care when validating this circuit, and would probably think about some kind of compensation scheme. – swineone Apr 18 '20 at 14:51
  • @swineone Yes, lowering the phase margin at low closed loop gains can cause peaking at the top end of the closed loop frequency response (frequency domain) and more overshoot in the time domain step response. If the phase margin is reduced further still then the peaking increases in amplitude and can turn into oscillation even with the input signal to the amp removed. –  Apr 18 '20 at 15:21
  • @James are you asking me a question or just saying something that should not be specifically addressed at me? The point of my answer was to show that there is a region of instability caused by adding more output capacitance but there is light at the end of the tunnel. – Andy aka Apr 18 '20 at 16:44
  • I'm asking why, in the third simulation down, you think it will oscillate when you haven't got the conditions for oscillation which would be loop gain = 1 (not>1) and 180 degs lag occurring simultaneously? Then in the 4th and 5th simulations you have virtually got the perfect conditions for oscillation and you think it won't oscillate. Or am I misunderstanding Barkhausen? I am quite happy to be corrected if I am getting things completely wrong! I can appreciate your underlying point though, that adding capacitance to the output reduces stability margins possibly to the point of oscillation. –  Apr 18 '20 at 18:44
  • The 3rd simulation has a phase shift of 180 degrees at a gain greater than unity. This means it will oscillate (and you are misled in your appreciation of conditions for oscillation in amplifiers). The loop gain for a sinewave oscillator has to be controlled to unity to keep sine wave amplitude stability. If the gain is greater than 1 then the output amplitude hits the rails of the op-amp. My underlying point is that if you keep adding capacitance you will reach a point of stability @James – Andy aka Apr 18 '20 at 19:24
  • 4th is borderline and 5th is not going to oscillate if we assume that there is 2 deg PM at gain = unity. I think your misunderstanding is based on sinewave oscillators. In other words if gain is equal to or greater than 1 and there is a frequency where the PM is zero, it will sing (if it can begin oscillating). – Andy aka Apr 18 '20 at 19:34
  • I've got it sussed now. Thanks for your help Andy. –  Apr 19 '20 at 19:42
2

If the opamp is not suited for that, driving a large capacitive load can indeed lead to instability problems if a compensation network is not implemented. The problem arises, when the capacitive load is high enough to cause a substantial delay (phase) between the output(feedback) and input signal, up to the point where regulation is very challenging.

At least at start up, the capacitor connected between the positive rail and the output/feedback path, provides a temporary "phase boost", by pre-charging the output capacitor directly with the supply rail. Meaning, that with the proper capacitive ratio \$(\frac{C_{22}}{C_{22}+C_{24}})\$ the output capacitor could theoretically be pre-charged with the stabilized voltage (\$1.5V\$) you are expecting to have in the non-inverting input.

vtolentino
  • 3,529
  • 1
  • 7
  • 17
2

The circuit specifies a particular part number of tantalum electrolytic capacitor on the output, which acts as a damping network (0.25 ohm ESR series with 47uF) at some frequencies, but I still would be suspicious this could oscillate under some conditions with only a small amplitude being visible, resulting in excessive power consumption and other subtle effects.

This looks more like something that was tinkered with until it sort-of worked than a carefully-designed robust circuit.

Spehro Pefhany
  • 376,485
  • 21
  • 320
  • 842
1

Andy already gave gave an excellent answer so I won't say more on the "oscillation" bit.

However I just want to add that having C22 is a mistake:

enter image description here

At high frequencies the output impedance of the opamp will be much higher than the impedance of the caps, so the output impedance of this circuit will be dominated by the caps. This means C22 allows any high frequency noise in VCCA power supply to leak directly into the output.

This is mitigated by C23/C24 which act as the lower leg of a capacitive divider with C22, so noise on VCCA will be attenuated accordingly. Still it is better to not use C22 at all.

Also C22,C34,C24 act as a decoupling cap for the VCCA rail. This means a nearby chip (like an ADC) which draws pulsed current from VCCA will draw some of its supply current from these capacitors, which is another way to add noise in the output.

bobflux
  • 70,433
  • 3
  • 83
  • 203
1

Andy-aka has done a good quantitative analysis but I wanted to take a more qualitative approach.

In general, attempting to apply negative feedback to a multi-stage amplifier where all stages have similar frequency response will result in oscilation.

The reason for this is the phase response, to a first approximation before the cutoff frequency a stage produces no phase shift and after the cutoff frequency it produces a phase shift of 90 degress. If all the stages have similar frequency response then the system encounters a large phase shift soon after the shared cutoff frequency. That large phase shift allows the aplifier to oscilate.

The fix for this is to make one stage have much lower bandwidth than the others. Generally in an op-amp one of the internal stages will have a compensation capacitor to deliberately reduce it's bandwidth. Sometimes the compensation capacitor is fitted externally to allow it's value to be tailored to the needs of the circuit.

Putting capacitance on the output reduces the bandwidth of the output stage, this brings it closer in bandwidth to the deliberately bandwidth-limited stage which can lead to oscilation.

But if you keep piling on the capacitance and hence reducing the bandwith of the output stage then things cross-over. The output stage eventually comes to dominate the frequency response and the second (originally first) break frequency is not reached until after the loop gain drops below one.

Peter Green