1

I am having problems calibrating my test fixture for the Agilent 4294A Impedance Analyzer. The open measurement always works, but with the short measurement I always get the error message "Bridge Unbalanced".

The same problem was described a few years ago (https://community.keysight.com/thread/4715), but never answered.

I changed all parameters like amplitude, frequency, bandwidth etc., but without success. Does anyone have an idea what the problem is or what you could do about it?

SamGibson
  • 17,231
  • 5
  • 37
  • 58
Charly
  • 650
  • 3
  • 14

2 Answers2

1

I found the problem. There was a broken fuse on the DC power source board A4. I renewed it (just a few cents) and now the device is working again.

Charly
  • 650
  • 3
  • 14
  • 1
    Excellent hint, thank you! Replacing the fuse cured our 4294A too that showed the "unbalanced bridge" issue. – Murmel Nov 09 '22 at 12:30
0

Ok, it's going to be an invalid setup/value for the short compensation.

Take a look at this answer at Keysight.

GB - AE7OO
  • 1,043
  • 7
  • 19