I am trying to determine if equipment I already have can be used to take a reading in mcd/m2, for a quality control test. I have a sensor which meets all the requirements listed in the standard test method except that it has a linearity error of <1% instead of linearity error ≤0.5%. Could a better calibration correct the linearity of this sensor so that it is within tolerance or do I need a better sensor?
https://www.intl-lighttech.com/products/ilt5000-researchlab-radiometer
https://www.intl-lighttech.com/sed033-broadband-silicon-detector
https://www.intl-lighttech.com/products/ilt5000-photometer-sed033yr-detector-luminancebrightness-measurement-system
https://www.intl-lighttech.com/products/ilt5000-sed033yw-research-lux-measurement-system
According to the mfg, this system satisfies all the following requirements except for linearity:
spectral error, f1' ≤5 %
UV response, u, ≤0.5 %;
resolution at least 0.1 mcd/m2;
resolution 1.0 lux;
linearity error, f3, ≤0.5 %
signal-to-noise ratio: at least 10:1 for all measurements.
EDIT: MORE INFORMATION
Data Mfg provided when asked about linearity:
Distance 1: 1 meter (lm/ft2) Distance 2: 2 meter (lm/ft2) Expected
Value % Deviation
2.42E+02 6.00E+01 6.050E+01 -0.8%
2.32E+01 5.75E+00 5.789E+00 -0.8%
2.20E+00 5.56E-01 5.509E-01 0.9%
2.60E-01 6.49E-02 6.510E-02 -0.3%
3.98E-02 9.86E-03 9.956E-03 -0.9%
4.66E-03 1.16E-03 1.164E-03 -0.5%
5.14E-04 1.28E-04 1.286E-04 -0.3%
Cal Values from this specific equipment:
No device found at COM port 3
Device found at COM port 4
Type help for a list of device commands
Type exit or quit to close console
getinfo
Base Serial Number = 00000000000000000
Vendor Serial Number = ILT500000000
Model Name = ILT5000
Friendly Name = NOT-DEFINED
Generation = 3
FW Version = 3.1.2.8
Dark mode (0=No,1=Factory,2=User) = 1
Rf Setting (0=Auto,1=Rf1,2=Rf2,3=Rf3,4=Rf4) = 0
Rf value R1 (kOhms) = 3
Rf value R2 (kOhms) = 1000
Rf value R3 (kOhms) = 10000
Rf value R4 (kOhms) = 10000000
Rf value in use (KOhms) = 10000000
eCal: Enabled
eCal values = 1.016e-12 4.965e-12 1.002e-11 5.044e-11 1.011e-10 4.963e-10 9.982e
-10 4.997e-09 1.000e-08 4.992e-08 9.991e-08 4.970e-07 9.946e-07 5.014e-06 9.994e
-06 5.007e-05 9.993e-05 5.002e-04 1.001e-03
Sample time (ms) = 2000
Auto Sample Time: Enabled
4-20mA mode (0=Auto/log, 1=min-max/linear, 2=manual) = 0
getvx1 = 0.081090
getvx17 = 1.381653
getvagc3 = 3.281837
Active voltage gain stage = 2
TIA voltage = 0.064558
getcurrent = 6.443e-12
Factory dark = R1 192 13 6 R2 172 47 65 R3 205 52 69 R4 17529 17299 14838
User dark = R1 160 0 0 R2 146 1 0 R3 159 1 0 R4 9272 9503 9558
Ambient level = 0.000e+00
Logging: Disabled
Wireless/Auxiliary Listening: Enabled
Peak Tracking: Disabled
Fast Integrate: Disabled
Multidrop: Disabled
Sensitivity: 1.475000e-09
Sorry if this question seems confused or is missing any needed info, I am just a tech, not an engineer.
Thanks.