I am trying to determine if equipment I already have can be used to take a reading in mcd/m2, for a quality control test. I have a sensor which meets all the requirements listed in the standard test method except that it has a linearity error of <1% instead of linearity error ≤0.5%. Could a better calibration correct the linearity of this sensor so that it is within tolerance or do I need a better sensor?

https://www.intl-lighttech.com/products/ilt5000-researchlab-radiometer
https://www.intl-lighttech.com/sed033-broadband-silicon-detector
https://www.intl-lighttech.com/products/ilt5000-photometer-sed033yr-detector-luminancebrightness-measurement-system
https://www.intl-lighttech.com/products/ilt5000-sed033yw-research-lux-measurement-system

According to the mfg, this system satisfies all the following requirements except for linearity:

spectral error, f1' ≤5 %
UV response, u, ≤0.5 %;
resolution at least 0.1 mcd/m2;
resolution 1.0 lux;
linearity error, f3, ≤0.5 %
signal-to-noise ratio: at least 10:1 for all measurements.

EDIT: MORE INFORMATION

Data Mfg provided when asked about linearity:

Distance 1: 1 meter (lm/ft2) Distance 2: 2 meter (lm/ft2) Expected Value % Deviation
2.42E+02 6.00E+01 6.050E+01 -0.8%
2.32E+01 5.75E+00 5.789E+00 -0.8%
2.20E+00 5.56E-01 5.509E-01 0.9%
2.60E-01 6.49E-02 6.510E-02 -0.3%
3.98E-02 9.86E-03 9.956E-03 -0.9%
4.66E-03 1.16E-03 1.164E-03 -0.5%
5.14E-04 1.28E-04 1.286E-04 -0.3%

Cal Values from this specific equipment:

No device found at COM port 3
Device found at COM port 4

Type help for a list of device commands
Type exit or quit to close console

getinfo
Base Serial Number = 00000000000000000
Vendor Serial Number = ILT500000000 Model Name = ILT5000
Friendly Name = NOT-DEFINED
Generation = 3
FW Version = 3.1.2.8
Dark mode (0=No,1=Factory,2=User) = 1
Rf Setting (0=Auto,1=Rf1,2=Rf2,3=Rf3,4=Rf4) = 0
Rf value R1 (kOhms) = 3
Rf value R2 (kOhms) = 1000
Rf value R3 (kOhms) = 10000
Rf value R4 (kOhms) = 10000000
Rf value in use (KOhms) = 10000000
eCal: Enabled
eCal values = 1.016e-12 4.965e-12 1.002e-11 5.044e-11 1.011e-10 4.963e-10 9.982e -10 4.997e-09 1.000e-08 4.992e-08 9.991e-08 4.970e-07 9.946e-07 5.014e-06 9.994e -06 5.007e-05 9.993e-05 5.002e-04 1.001e-03
Sample time (ms) = 2000
Auto Sample Time: Enabled
4-20mA mode (0=Auto/log, 1=min-max/linear, 2=manual) = 0
getvx1 = 0.081090
getvx17 = 1.381653
getvagc3 = 3.281837
Active voltage gain stage = 2
TIA voltage = 0.064558
getcurrent = 6.443e-12
Factory dark = R1 192 13 6 R2 172 47 65 R3 205 52 69 R4 17529 17299 14838
User dark = R1 160 0 0 R2 146 1 0 R3 159 1 0 R4 9272 9503 9558
Ambient level = 0.000e+00
Logging: Disabled
Wireless/Auxiliary Listening: Enabled
Peak Tracking: Disabled
Fast Integrate: Disabled
Multidrop: Disabled
Sensitivity: 1.475000e-09

Sorry if this question seems confused or is missing any needed info, I am just a tech, not an engineer.

Thanks.

str3r
  • 23
  • 4
  • 1
    What photoreceptor, what circuit? Please provide information that we need to answer the question. – Voltage Spike Sep 30 '19 at 21:20
  • Are you asking for the model of the photoreceptor? It's not a measurement of a circuit, it's a measurement of a luminescent material. – str3r Sep 30 '19 at 21:23
  • 1
    And what circuit are you using the photo receptor in? that matters for linearity. Also edit your question with the values that you provided in the comment – Voltage Spike Sep 30 '19 at 21:24
  • Values added to question. I'm not building a circuit, it's a calibrated, self-contained system. I may be in the wrong place but I couldn't find a photonics or optics stackexchange. – str3r Sep 30 '19 at 21:31
  • But to calibrate you need an even more linear device...and this would not be a calibration as much as a linearization which means you need processing hardware. – DKNguyen Oct 01 '19 at 01:51
  • yes it can ... you can generate a lookup table that correlates the reading value to the physical value ... the table does not have to be "fine grained" because you could interpolate the values in between the lookup table entries – jsotola Oct 01 '19 at 03:25
  • _"I have a sensor which meets all the requirements listed in the standard test method except that it has a linearity error of <1% instead of linearity error ≤0.5%"_ which sensor is that? None of the products in your links appear to specify a linearity error. – Bruce Abbott Oct 01 '19 at 04:04
  • It's important to distinguish between accuracy, resolution and also linearity and repeatability. All these terms need to be well defined to be useful, with the latter two being (perhaps) potentially 'woollier'. It is essential to specify accuracy and resolution. Can you cite data sheet entries for these figures and for the one that you mention? – Russell McMahon Oct 01 '19 at 05:30
  • ILT5000 datasheet [here](https://www.intl-lighttech.com/sites/default/files/pdf//datasheet/ILT5000-Data-Sheet-Final.pdf) - it is utterly and outsandingly amazing that nowhere do they give ANY performance limitations in terms of number of bits, bits of accuracy (or bits of anything), % accuracy (or % of anything) . Where did you find the 1% linearity figure that you mention? – Russell McMahon Oct 01 '19 at 05:36
  • @Bruce - ILT5000 datasheet [here](https://www.intl-lighttech.com/sites/default/files/pdf//datasheet/ILT5000-Data-Sheet-Final.pdf) - no real 'performance' specs at all it seems. – Russell McMahon Oct 01 '19 at 05:37
  • @RussellMcMahon I got this info from a conversation with the mfg, but they won't let me speak to their engineer. – str3r Oct 01 '19 at 17:03

2 Answers2

0

The performance is the performance. No amount of "calibrating" is going to take a non-linear sensor and make it more linear.

That said, you could linearize around your region of interest, characterize the performance and fit to a higher-order function, build a lookup table, ..., to provide a more accurate relationship between input and output.

Scott Seidman
  • 29,274
  • 4
  • 44
  • 109
  • Isn't correction to the CIE photopic curve already part of the calibration of photometers in the visible range? If so, why does it matter how linear they are to begin with? – str3r Oct 01 '19 at 17:39
  • The CIE photopic curve specifies the relative response to different frequencies. It is indirectly related to linearity in that the less linear the device the harder it may be to accurately correct the photopic curve at different intensities. But silicon photosensors generally have very good linearity, so I bet they only measured it at one light level (and the CIE photopic curve itself is not 'accurate' at low light intensities anyway). – Bruce Abbott Oct 01 '19 at 20:40
  • Had to stare at this for a little while but I think get it - CIE curve corrects frequency response, linearity has to do with intensity? If so, makes more sense now that they would want a device with very good linearity for this measurement of something relatively dim. – str3r Oct 01 '19 at 21:30
  • Their meaning of linearity is shown in the added data. They measure a light level at a distance of 1 & 2 metres. They predict the 2m level to be 25% of the 1m value and calculate a %age. Specs for eg accuracy are usually given as a % of full scale. As this device is wholly autoranging across 10 decades that's not doable, but the figures given are "as good as you need for a;most any real world task". Some may need better - but in this test the combination of distance measurement , illumination field linearity, ... need to be well under 0.5% to allow actual meter response to be determined. ... – Russell McMahon Oct 02 '19 at 10:18
  • eg to get 0.1% accuracy in light reading when moving from 1m to 2m the 2m distance needs to be correct to within 1mm and the 1m distance to 0.5mm. This is from light source to sensor integrating window. One has to decide (unless data not in datasheet or in some other document is available (or I missed it, as ever)) where the correct distance point is on the input surface. ie IF the system is as good as shown then one would have to try exceptionally hard indeed to set it up so as to ensure readings which were able to be as good as the meter specifications. So - it will do the job. – Russell McMahon Oct 02 '19 at 10:27
  • [Manual](https://www.intl-lighttech.com/sites/default/files/downloads/datalight_iii_manual.pdf) & [datasheet](https://www.intl-lighttech.com/sites/default/files/pdf//datasheet/ILT5000-Data-Sheet-Final.pdf) – Russell McMahon Oct 02 '19 at 10:28
  • 1
    @Scott What you say. BUT - see my answer re making measurements that make the meter specs relevant. – Russell McMahon Oct 02 '19 at 10:54
0

Manual here: &
Datasheet here:

If the information provided by the manufacturer directly and in documentation is taken at face value the the meter will meet your requirement.

The provided linearity values vary slightly above and below the 0.5% level BUT it would be almost impossible to use the meter + sensor system in such a manner that the accuracy predicted by the linearity was reached by the experimental setup. That is - the meter specs are slightly better than the variations that a careful user is likely to achieve in practice experimentally.

This is good :-).

If an instrument has a 0.5% linearity, for this to matter, factors relating to the setup and measurement need to be substantially smaller than the errors contributed by the meter & sensor.

As a "metaphor":

  • If a vehicles moving-needle speedometer has an accuracy (in this case rather than linearity) of 1% at 100 kph and a driver wishes to measure a distance (say of ABOUT 600 metres) by driving at 100 kph and measuring the time taken, then the ability to hold the vehicle at a steady 100 kph indicated and the accuracy of the timing are both relevant factors in how 'good' the measurement will be. Based on my personal experience, on a level road it would be hard to hold a vehicle at better than +/- 1 to 2 kilometres per hour of 100 kph for say 1 minute, and timing the time to cover the distance with a manually controlled stopwatch may introduce errors of +/- 0.1 to say 0.3 seconds.
    SO speed error is around 1.5% and time error around 0.25/60 ~= 0.5% so achievable accuracy would be in the +/- 2% range - which while not "swamping" the 1% speedometer accuracy, makes it a lesser factor than may otherwise be expected.

Back to the light meter:

To make use of a meter's 0.5% linearity a setup accuracy of around say 0.1% is desirable.
To get 0.1% accuracy in light reading when moving from 1m to 2m the 2m distance needs to be correct to within 1mm and the 1m distance to 0.5mm. (The smaller than 0.1% distances are due to the inverse square law effect).

These distances are from the light source to the sensor integrating window. One has to decide (unless data not in datasheet or in some other document is available (or I missed it, as ever)) where the correct distance point is on the input surface.

Also, The light source needs to be uniform across the area of the integrating window in each case (or of effectively equal uniformity - don't start :-)).

IF the system is as good as shown then one would have to try exceptionally hard indeed to set it up so as to ensure readings which were able to be as good as the meter specifications. So - it will do the job. https://www.intl-lighttech.com/sites/default/files/downloads/datalight_iii_manual.pdf) Manual & datasheet

"Input optics" - worth looking through.
Note that they say " ... A well defined measurement plane and approximate point source are required for valid interpolations...."

and, depending what you are measuring, it may be worth noting: interesting

"In radiance measurements (W/cm²/sr), where a detector with a narrow viewing angle looks at a much larger, uniform area source, the area of the source that the detector sees increases as the detector moves farther away, counteracting the effects of the inverse square law and making the reading independent of distance."

Russell McMahon
  • 147,325
  • 18
  • 210
  • 386
  • Thank you! Definitely worth nothing as I have to select measurement distance based on a minimum sample area. – str3r Oct 03 '19 at 22:14