I recently bought a basic digital oscilloscope and wanted to power my oscilloscope from 600VA computer UPS.
I have two questions:
  1. Can the UPS give isolation from main supply? (As an alternative to an isolation transformer.)
  2. The output of the computer UPS waveform is "simulated sine wave." Will it affect the performance of the oscilloscope?
UPS datasheet
JRE
  • 67,678
  • 179
masternone
  • 290
    You should not be floating your scope. Float your device under test, not your scope. Floating your scope makes things like all the BNC exposed connectors, probe ground clips, and the voltages right behind the knobs and buttons on your scope float all the way up to whatever you are testing which is dangerous. – DKNguyen Jul 02 '19 at 04:58
    First, although most power supplies are not officially certified to use with modified sine wave as power source, but often the majority of switched-mode PSU should work without problems, including oscilloscopes, so yes, you can use a UPS to provide backup power to your scope. Next, if you UNPLUG your UPS and run it on batteries, the inverter is no longer ground-referenced, and it does provide isolation in a sense, but it's dangerous and YOU SHOULDN'T DO IT, see https://electronics.stackexchange.com/questions/73991/why-do-we-need-an-isolation-transformer-to-connect-an-oscilloscope/73994#73994 – 比尔盖子 Jul 02 '19 at 05:30
  • after long surfing about powering oscilloscope, i wrongly understand to isolate oscilloscope instead of testing unit, thank you all – masternone Jul 02 '19 at 06:35
1- does computer UPS give isolation from main supply? ( like alternative to isolation transformer)
No, "UPS 600VA" it is not isolating from main supply
2- output of the computer UPS waveform is "simulated sine wave", does if affect performance of oscilloscope
No it doesn't affect the performance because it converts to DC inside the oscilloscope.
A simulated sinewave is basically a pulse width modulated squarewave, with a proper output and/or input filter i would not expect any issues.
shall i use computer UPS for oscilloscope
No, I don't see any reason for this unless you live in a place with a lot of short-term power outages.
Daniel K
  • 699
  • A UPS could be used to provide isolation from the supply. Plug in the UPS for a few hours to charge the battery inside. Then unplug it from the supply, and plug the oscilloscope in to the UPS. – Simon B Jul 02 '19 at 09:45
  • I think you are wrong. When you unplug UPS from the mains it is pretty much isolated from the mains. – Chupacabras Jul 02 '19 at 11:47
  • This is not the purpose of a UPS @Chupacabras Never float a oscilloscope: https://electronics.stackexchange.com/questions/73991/why-do-we-need-an-isolation-transformer-to-connect-an-oscilloscope/73994#73994 – Daniel K Jul 02 '19 at 13:22
  • I didn't say it is a purpose of any UPS. I just commented on your statement which is false. Oscilloscope connected to UPS (which is disconnected from the mains) is isolated from mains. No matter whether it is recommended or not. – Chupacabras Jul 02 '19 at 14:42
  • For sure it is isolated but nobody will operate an oscilloscope like that. Therefore I did not consider this case. – Daniel K Jul 02 '19 at 15:53
  • @DanielK Your statement is still false. Doesn't matter you didn't consider something. – Chupacabras Jul 03 '19 at 06:36
I've seen quite a few pieces of test equipment used for years on a UPS. Including oscilloscopes. As far as I know, there hasn't been any issues. The ground passes through the UPS so the test equipment is still at earth ground. Sometimes the neutral passes through only when the UPS is powered from the mains, but sometimes it is isolated through a transformer.
I myself have a whole house battery backup system and many times the power has been out when I've been working on something, and I never knew. It is a very pure sine wave.
Only if the UPS provides isolation (and it would have an isolation barrier internally) Some UPS's convert AC to DC and then back to AC with switchers.
It really depends on the setup, but if the ground of the oscilloscope is isolated from earth, then it's most likely floating. Whatever your measuring would need to be plugged into the UPS or there really wouldn't be a reference for the oscilloscope ground.
Whatever your measuring also needs to be grounded to the oscilloscope.
This is also not a good way to measure any high voltages because the ground of the oscilloscope would need to be connected to the high voltage device and that could create a saftey hazard.
The best way to make isolated measurements is with a differential probe because it has two terminals and determines the voltage between them
Voltage Spike
  • 75,799
  • 80
  • 208