I would like to build a reference thermometer for checking and maybe calibrating other sensors and I was planning to use 1/10 DIN RTD probe in order to do that. You can get such probes for ~$100 and the error is ≤±0.07C in -50°C - 50°C range, which sounds great. I'm not sure, however, if I can use it efficiently without deep electronics knowledge or spending a few hundred dollars on a reference thermometer like Fluke 1523.
The easiest (and pretty cheap) way to use RTDs is MAX31865 (IC to automatically convert RTD resistance to temperature readings), but the only note about accuracy that I can find on it is this line from the datasheet "Total Accuracy Over All Operating Conditions: 0.5°C (0.05% of Full Scale) max". 0.5°C full scale doesn't sound bad, but I'm not really sure what that means for lower temperature ranges, for example: how much the error of the RTD will increase in the 0-100°C range when using MAX31865. Without that knowledge I also don't know how much do I gain by using 1/10DIN instead of Class B probe.
So my general question is: what's the easiest way to accurately read precision RTD probes with a limited budget (like $50-$100 not including the probe)?
Piotr Sarnacki
  • 236
    It is a shame a not minor number of IC has not a very precise resolution for measuring sensors...... – Brethlosze Dec 31 '16 at 12:31
  • Personally I have had a lot of success with the AD7709 and a 5ppm (or better) temperature coefficient reference resistor. – Matthew Dec 31 '16 at 20:47
  • @Matthew thanks for the pointer, I haven't seen this one before. The only problem is that here, too, I can't really tell how accurate it would be and there's virtually zero information on using AD7709 with RTDs on the internet other than 2 example schematics in the datasheet. In that context I would probably prefer MAX31865 as at least you can get it ready to go with a breakout board and detailed instructions on RTDs. – Piotr Sarnacki Jan 01 '17 at 10:07
  • Link to a datasheet for this MAX... thing. You haven't even said if it's a A/D, a specialized IC for RTDs, etc. – Olin Lathrop Jan 02 '17 at 22:48
  • @OlinLathrop sorry, I added the link to the datasheet just now. In the meantime I also found one more IC that could help: [AD1247/AD1248](http://www.ti.com/lit/ds/symlink/ads1248.pdf). While it seems similar to AD7709 which was mentioned above, there is much more info on using AD1247/8 with RTDs, including the [application note](http://www.ti.com/lit/an/sbaa180b/sbaa180b.pdf) and support from TI, including the [answered question about accuracy](https://e2e.ti.com/support/data_converters/precision_data_converters/f/73/t/220458). – Piotr Sarnacki Jan 03 '17 at 00:00
  • What accuracy do you need? What resolution? – Tony Stewart EE75 Jan 04 '17 at 13:25
  • you cannot engineer something without specs ! i.e. What is your acceptance criteria: offset and gain error, range and budget? then a good answer is possible. – Tony Stewart EE75 Jan 04 '17 at 15:43
  • @TonyStewart.EEsince'75 thanks for pointing it out. Ideally I would want a total accuracy of ±0.15C in 0-100C range with 1/10 DIN probe. I think that would let me calibrate other sensors with ±0.3C accuracy. When it comes to resolution I think 0.01C or better would be nice. I'm not really sure about offset and gain error. My budget is ideally $50, but I could spend more if that's too low. But all that is not really a hard requirement, I'm just a hobbyist. If I can do better in this price range, then great! If it's unrealistic, then so be it. – Piotr Sarnacki Jan 04 '17 at 18:23
The MAX31865 seems like a very good choice. The datasheet does give more detailed accuracy specifications—the first page of a datasheet is generally just marketing material.
On page 3, we have some specifications of the ADC: full-scale error typically ±1 LSB, integral nonlinearity typically ±1 LSB, and offset error at most ±3 LSB. Therefore, the output of the ADC will typically be within 4 least-significant bits of the correct value. Since it's a 15-bit ADC, that's an error of \$\frac{4}{2^{15}}\$ or about 0.013%. Since the resistance of an RTD is roughly linear with temperature (at 2.73 K/Ω for a PT100 RTD), an error of 0.013% at 273 K corresponds to a temperature error of 0.013% * 273 K = 0.036 K. At a higher temperature, the absolute error would be proportionately larger.
We also have some graphs of accuracy in the datasheet, at the bottom page 6. These give a more detailed picture, and we can see that our previous estimate of a typical error of 0.036 K is not too far off, though the absolute error does not seem to grow linearly with resistance.
Overall, the 0.5 K worst-case error from page 1 of the datasheet seems very conservative. The actual error is likely to be no more than the error inherent to the RTD, if you do indeed go with a 1/10 DIN RTD, which has an error of at most ±0.07 K from -60 to 50 degrees Celsius.
If you follow the datasheet recommendations, though, you'll introduce another source of error: self-heating. The datasheet suggests a 400 Ω reference resistor for a PT100, which results in 4 mA through the RTD (due to the 2 V bias generated). This is about an order of magnitude above the recommended current for a PT100, so depending on what RTD probe you pick, you may want to use a reference resistor of about 5 kΩ instead.
The advantage of a smaller reference resistor (and therefore a higher current is better noise-immunity) so there is a tradeoff. For use as a reference thermometer, you can afford to average the noise out over a very long time, so a large resistor makes sense. In an industrial application, you might have a lot of noise, but also a very large measurement sample that can absorb the generated heat effectively, so there the smaller resistor would make sense.
Interestingly, regardless of the choice for reference resistor, the tolerance of it is very important, as it directly contributes to the final error. Using an 0.1% resistor means that you can never do better than 0.1% error (0.273 K error at 0 degrees Celsius), so you may want to splurge on an 0.01% resistor.
Abe Karplus
  • 2,611
  • Thanks, it explains a lot! It's interesting on what you wrote about the reference resistor. [Adafruit's breakout board](https://www.adafruit.com/product/3328) uses 430Ω 0.1% resistor. – Piotr Sarnacki Jan 03 '17 at 14:39
  • @PiotrSarnacki I've added some more details about choosing reference resistors. – Abe Karplus Jan 03 '17 at 18:33
  • Thanks! In [an article that I found](http://electronicdesign.com/embedded/eliminate-rtd-self-heating-errors) they claim that self heating effect can take a few minutes to settle, so maybe as long as you don't take measurements continuously it would be fine. But even then some RTDs that have current specified at 1mA max. – Piotr Sarnacki Jan 03 '17 at 23:27
  • One more thing that I noticed is that MAX31865 gives you a ratio of RTD resistance to reference resistance, so one way to avoid the error would be to measure the actual resistance of a reference resistor and use that in calculations - that seems OK for a one-time project. – Piotr Sarnacki Jan 03 '17 at 23:30
  • @PiotrSarnacki Sure, if you can measure resistance to better than 0.1%. Personally, I'd spend the $1.09 (in quantity 1) for an 0.01% reference resistor. – Abe Karplus Jan 03 '17 at 23:52
Industry standard for checking RTD Probes is a two set process.
Step 1: remove the RTD probe from the system and insert it into a dry block where you can set different temperature ranges. As the probe is run through temperature ranges it is read using a certified Fluke calibrator.
Step 2: Using a certified Fluke calibrator source and simulate the RTD signal into the process wiring where the RTD was installed. This is simulated across the same temperature range as the dry block.
I've considered something similar to what you propose. I would order a RTD Probe and meter combination from omega.com and having the set certified annually for a given temperature range. I have also considered buying precision resistors across the RTD range that correlates to the temperature ranges my process is exposed to. For me this may be a quicker way to accomplish Step 2.
Tinkerer
  • 1,033