I operate an MEG system (extremely sensitive magnetic sensors for measuring brain activity) and I recently found that the DC power source that feeds into our magnetic shielded room (power used for moving the MEG/chair/bed - thus required), produces 60 Hz line noise detected by the sensors. I'll detail my understanding the sources of this noise, but my question basically is: what kind of supply or what specific features should I look for to avoid this noise?
  1. Noise from currents in wires. The current supply is a typical cheap switched-mode floating DC supply, 24V 0.5A. I did a few tests to confirm that although the difference between + and - looks "clean" and constant, each wire has a large voltage oscillation w.r.t. ground. And this "voltage" seems to be the result of a current source, so even with no load, there will be an oscillation of charge density in the wires, enough to generate a magnetic field detected by our system. (In the kOhm to MOhm range between + and G, I get on the order of 0.1 mA peak current.)
  2. Noise from ground currents. Unfortunately, if I tie the negative to ground, of course the oscillation w.r.t. ground is gone, and that mostly gets rid of the detected noise, but the "offending" current is not eliminated, instead going through the ground. That also causes some noise, probably primarily because our shielded room is itself grounded. Worth mentioning that this noise is much smaller.
Edit: Price is not a primary concern. I realize a well designed low-noise supply would likely be much more expensive.
zorgkang
  • 93
    Possibly a battery. Something like a car battery or wheelchair battery given the loads you're driving. –  Oct 24 '16 at 16:14
  • @Brian Thanks, good point. That definitely deserves consideration. But might be somewhat less convenient than a supply I can just "plug and forget". Curious about other options. – zorgkang Oct 24 '16 at 16:20
  • A real big filter coil outside the room. And staying clear from earth with all conductors. But I favor the battery solution. – Janka Oct 24 '16 at 18:12
  • Linear or switch mode power supply? Is the noise radiated or conducted? Since your system is a sensitive magnetometer it could very well be that you are seeing radiated 60 Hz B fields from the transformers in the linear power supplies. – George Herold Oct 24 '16 at 18:57
  • @George Switched-mode as I stated. And I tested with the supply powered, but disconnected from the (shielded) extension cable that goes into the room and the noise disappears. So it's from currents in the cables. – zorgkang Oct 24 '16 at 19:17
    Can you measure the noise (e.g with another wire next to the sensors) and subtract it from the signal? – Andrew Morton Oct 24 '16 at 19:45
  • @zorgkang, how much voltage/ current? I might try Marcus's idea of adding a linear regulator after the SMPS. – George Herold Oct 25 '16 at 15:24
  • I think Mag sensors to brain is a **fruitless exercise** unless the designer has complete understanding of EMC, considering dielectric currents and exposed loop is so small. There are no high current conducting wires in the brain exposed. Only E fields can be measured not H fields, unless excited externally. I would recommend a Lindgren (brand) cage for starters – Tony Stewart EE75 Oct 25 '16 at 19:05
  • A friend of mine designed the world's first mobile MRI machines using all plastic hydraulics. He was an expert on RF and EMC. The MRI used to implode all raster scans of all the PC CRT's in the NRC building when it was operating. (before LCD's became popular) – Tony Stewart EE75 Oct 25 '16 at 19:12
  • I'm not saying MAG sensing is impossible, but you must know everything about EMC to trace such low level H fields safely with zero 60Hz... and we cannot do a complete technology dump in one page. – Tony Stewart EE75 Oct 25 '16 at 19:14
  • To measure E field CM noise attach a sheet of Aluminum foil to a 10M scope probe and measure it. The coil wrap ought to be as large as a human body which acts like an antenna for E field.... For H field RF short the probe to itself. – Tony Stewart EE75 Oct 25 '16 at 19:17
  • This is a commercial medical device for decades. And has been my job for 8 years to operate it. I'm trying to reduce 60 Hz noise, but the system works fine otherwise and how it works is completely besides this post's scope. People have just been filtering out 60 Hz, I'm trying to improve this slightly that's all. – zorgkang Oct 25 '16 at 19:31
  • @Andrew Yes there are other approaches to reduce noise in MEG. What you mention is part of the system. But removing it at the source when possible is usually the best. – zorgkang Oct 25 '16 at 19:45
If you really got a cheap switching power supply, then most of the noise you're seeing is probably not due to a lack of clean regulation, but because the AC lines couple back into your supply lines after the switch mode supply; your observations under "1." match that. It's really hard to avoid that – only a very dedicated device design and excellent measurement shielding can elimate that.
However, there's a couple of things that are often done to overcome this:
  1. Make sure you don't have any power lines running into your measurement chambers. That especially applies to ceiling lights etc.
  2. Make your ground better. If your ground level starts oscillating when you connect your supply ground to it, you simply have a high impedance w.r.t. to "earth".
Generally, if you want to build a somewhat efficient, cheap, low-noise power supply, you're often in for a Switching-Mode Power supply, like the cheap one you're using, followed by a relatively large capacitor that guarantees that load changes and slight fluctuations don't matter that much, followed by a low-ESR capacitor (typically, ceramic), followed by a linear regulator, followed, again, by a buffer capacitor. That way, you can get the "non-energy-burning, fast-to-adapt-to-load-changes" behaviour of the SMPS, and the low-noise properties of a good linear regulator (it might be worth looking a minimal bit further than the usual, ancient, 780X and LM317 regulators – there are a few more modern linear regulators on the market, and some of them have better noise immunity).
Marcus Müller
  • 88,280
  • 131
  • 237
  • To my knowledge, the state of the art in low-noise linear regulators is the LT3042 / LT3045. Note that careful attention must be given to PCB layout, connector, and shielding techniques to get maximum performance from a regulator when noise is this critical. – user49628 Oct 24 '16 at 18:57
  • Your point 1 is well taken care of. Regarding the ground, we do have a "somewhat dedicated" ground for our lab, but not fully "separate" as recommended by the manufacturer. I presume the ground quality was tested however when the system was installed in 2010. I don't know how I could possibly determine if my ground level changes without another ground reference. Regarding your last paragraph, to my limited knowledge this all sounds like it would reduce fluctuations of + w.r.t -, not "common mode" fluctuations w.r.t. ground. Can you clarify? – zorgkang Oct 24 '16 at 19:30
The "offending" current that you mention is a leakage current inside your switching power supply. This parasitic current comes from capacitive coupling between primary and secondary winding of the high-frequency isolation transformer, and creates parasitic currents at AC mains frequency. There could be a smaller (intentional) DC coupling, to avoid total isolation from earth ground. This AC leakage will be on non-switcher PS as well, since there will be still a transformer in between. It is not possible to avoid this coupling, one can only reduce it by making special isolation transformers with grounded shield between the windings at the expense of efficiency. Funny, I just looked into specifications for so-called Medical Power supplies, they list the approved leakage at 0.5mA, five times bigger than you have measured.
To reduce the effect of leakage currents, you need to design a good grounding scheme for the entire setup, to separate return currents from power electronics from grounds on sensitive instruments, and to return the leakage current into ground before it enters your shielded room. This is always a challenging problem. Good luck.
Ale..chenski
  • 38,845
  • 38
  • 103
I am going to take a slightly different approach to this.
Firstly, are you sure the pickup is magnetic? If so, then there are essentially two things you need to do:
  1. Connect one side of the DC supply to the rooms shielding right where the cables pass into the room (Both legs of the supply cable must pass thru the same hole in the room shielding), this ensures that circulating ground currents due to the EMC cap in the supply stay outside the room.
  2. All the internal wiring inside the room for stuff must be tightly twisted, this will ensure substantial field cancellation. Worst case use a group of four wires twisted around a common centroid (Used in microphone circuits and known as "Starquad" in that application), wiring is by paralleling opposite conductors, and it gets you maybe another 10 or 20dB of suppression.
If the pickup is E field rather then H, then screening the cables (and bonding the screen to the room shielding is an easy fix).
To answer the question asked, a linear supply built with a transformer having a grounded interwinding screen will be almost as quiet as a battery.
You might find that opening up your switcher and removing the cap across the isolation barrier, and then grounding one leg of the output to the room screening is useful, the supply will no longer meet EMC, but that may or may not matter to you.
Good luck, this stuff can be a bear to track down.
Dan Mills
  • 17,266
  • 38
  • Thanks. Yes these are magnetic sensors, it has to be. 1. is pretty much what I tried in my point 2. As I mentioned, it does work, though I still see something slightly more than when not powered, thus I suspect ground currents in the room shielding could be the cause. Regarding twisted pairs, that doesn't suppress common mode signals though, right? I will look into the "screened" linear supply you describe. Although I much appreciate understanding the inner workings, which is why I'm here, I'm not prepared at this point to make/modify a supply myself. – zorgkang Oct 25 '16 at 17:06
  • Common mode cannot be magnetic, unless there is some way other then your tight loop for current to flow (Possibly capacitive coupling in one of the loads to the room structure?), in which case it is not common mode any more. A suitably high value low frequency common mode choke may help, by forcing the currents in both legs to be equal, maybe take a 50Hz transformer having two secondaries and use the two secondary windings as a common mode choke with the primary disconnected. We used to wind common mode chokes out of RF coax on old transformer cores, quite effective. – Dan Mills Oct 25 '16 at 17:28
  • I'm saying the common-mode currents in the straight DC supply cable in my shielded room generate a magnetic field detected by my sensors. Not sure what you mean by "my tight loop". I was under the impression that a choke would not be effective at 60 Hz, or would need to be too big? What type of transformer are you thinking of? (Keep in mind I'm hoping for minimal tinkering.) – zorgkang Oct 25 '16 at 18:13
  • You must measure **conducted noise** using a scope probe without tip & ground clip using tip/barrel THEN **radiated noise** with ground clip shorted to probe clip and use like an antenna, to determine where EMI comes from, then reduce CM voltage by any means. Try to see shunt effects of differential L-C to gnd. This is a starting point for tests on EMI. Most likely 60Hz is due to unbalanced CM impedance 0.000?% to ground and excessive CM voltage. SMPS are only isolated at low frequency – Tony Stewart EE75 Oct 25 '16 at 19:01
  • @zorgkang you can eliminate **all your problems** with a suitable matched CM choke (CLC or "pi" diff choke) ( into Diff amp using STP wire.) {f matched means permeability for low or high frequency} – Tony Stewart EE75 May 20 '17 at 00:42
If you want lowest noise and are not too worried about cost ,size ,and losses then the linear supply is the best .They still make them at a price for lab use .If you want to make one yourself there is less to go wrong than with a switcher .If you still want to do a switcher then look at a soft switching topology or some resonant scheme .I did some Apples with Apples conducted EMC tests and found that my soft switch resonant circuit was no better around 150KHz but was much much better in the MHz range .The conducted EMC plot fell off fast and sudden with frequency like the Cliffs of Dover while the orthodox hard switched PWM was like a rolling hill .This implies that some switching loss reduction scheme will help you for radiated EMC .Remember that this is off course not as quiet as a linear ,and also note that many chinese power supplies have fraudulent EMC compliance stickers .Just test one and see!
Autistic
  • 14,235
  • 65